Coaxial Calibration



Vectorial VNA Calibration
Once all the required hardware are listed, click ‘Next’.
Define the frequency list and the VNA input and output sources power levels to use during the calibration. These power levels should not exceed the maximum power supported by the calibration kit, yet the level should be sufficient to provide noiseless and detectable signals at the VNA receivers even when adding attenuators to keep the receivers safe during the DUT characterization.

In the example above, the sweep type is frequency sweep starting from 1GHz to 3GHz with 50MHz step. In the example below, DUT has 20W output power and 30dB gain amplifier. The calibration will be performed with the RF source of the VNA set to 0dBm at the input and 10 dBm at the output. Remember that there will be a 30dB attenuator (input) and a 60dB attenuator (output) connected in front of the VNA receivers. To define the appropriated attenuator values on each receivers, see 2) Power Budget.
Once the frequency list and the VNA power level have been set, click ‘Next’.


When the calibration on the VNA interface is finished, go back to IQSTAR and press S-Parameters calibration Checkto validate your vectorial calibration or save a 'Thru.s2p' file. Learn more see Validate Calibration.
When the calibration on the VNA interface is achieved go back to the IQSTAR interface and press ‘Next' to continue.
Absolute Power Calibration
The power sensor can be eventually zeroed and calibrated before calibrating the absolute power.

If not, skip this step by clicking ‘Next’.
Define the power level Praw Offset used for the absolute power calibration step .

This step allows calibrating the VNA receivers for absolute power measurements using a power sensor as a standard. Setting an appropriate power level for the calibration of the receivers is the key for accurate calibration.

- Basic mode (“No power optimization done”)
In this case, the absolute power error term will be determined on a single power level for each frequency. This calibration method is used when the components of the bench (coupler, attenuator, …) are well-known and already sized to achieve the measurements in the linear range of the receivers.
Set the power level at which you desire to calibrate the receivers. The power meter measurements are compared with the receiver measurements in order to determine the absolute power error term coefficient. Using this error term, the absolute power of the RAW data measured by the receiver is corrected. There are several reasons for which you might change these settings:
- The power sensor that is used may have power limitations (typically +20 dBm). However, best accuracy is often close to 0 dBm
- If an external component is used in the measurement path, you
MUST adjust this RF source power setting to compensate for
these components. For example, if you want 0 dBm at the
calibration plane and you have a 30dB amp in the path, you
will need to set -30 dBm for this calibration.
Note: If a Driver Amplifier is set in the Schematic, the power level filled will take into account the Driver Amplifier offset
Note: In the example above, the power meter reading is not constant over frequency. This is due to the dispersive behavior of each passive component (couplers, attenuators, driver amplifier, …). In some configurations, this dispersive behavior can lead the power meter to change its measurement range adding uncertainties to the calibration process. In this case, the advanced mode is advised. - Advanced mode (“Power optimization done”)
In this case, the absolute power error term will be determined on a power range or on one optimized power level for each frequency. This method is longer than in the basic mode, but it will give also an estimation of the linearity range of the receiver. This calibration method is basically used when the test set bench (coupler, attenuator, …) is not well-known. Therefore, the receiver’s linearity range needs to be checked to evaluate the calibration accuracy.
During this process, the power level of the RF source will be optimized to reach a target value (Ptarget power meter) measured on the power sensor (at the input calibration plane). Then a power sweep will be done from Ptarget power sensor to [Ptarget power sensor– (1dB x Power Count)] in the calibration plane
- Ptarget Power Meter: corresponds to the maximum power level which has to be measured by the power sensor during the power sweep. Note that this level can’t exceed the power sensor limit
- Power count: corresponds to the number of point
of the power sweep. Note that the sweep step is set to 1dB Note: If the Power Count is set to 1, the optimization will be done only at one power level for each frequency.
- Praw Offset (initial) is the initial power
level used during the first iteration of the optimization algorithm.
Note that this power level is linked to the 'Minimum
Power' level set in the RF Source settings. There is a reason that you might change these settings:
- The power sensor that is used may have power limitations (typically +20 dBm). Therefore, the first power level used during the first iteration of the optimization process has to be lower than this limitation to avoid any damages.
Below is an example, where the 'Power count' is set to 1 in order to compensate the dispersive behavior of passive components.
Contrary to the 'Basic' mode, the power level measured by the power meter remains constant over the frequency range, avoiding the change of measurement range of the power meter.
Below is an example, where the 'Power count' is set to 30 in order to check the linearity range of the receivers.Note: Checking the linearity of the receivers can be a long process, it is advised to do it only on some representative frequencies.
In this example, the Ptarget power meter value is set to 0dBm, the algorithm applies a power level of Praw Offset (initial) (-30dBm) as a first iteration . During this first iteration, the power sensor returns a value of -34 dBm. In a second step, RF power source level will be adjusted (+4dBm corresponding to -30dBm+34dBm) to reach 0dBm on the power meter. Finally, a power sweep between Ptarget power meter (0dBm) and (Ptarget power meter– 1dB x Power Count) (-30dBm) will be done at each frequency in the calibration plane.
During the calibration process, the RF source level will increase until the power sensor reads the Ptarget power level expected. The process is repeated for each frequency defined in the frequency list (starting from the highest frequency). Once the calibration is completed the error terms are downloaded and recorded in the Calibrations tab. A comment is automatically added but it can be modified. This calibration will be used by default for the measurement unless the box Use this calibration is unchecked.
It is also possible to continue the calibration process by doing a ‘Raw Power Calibration’.Quick checks can be performed to verify the accuracy of this calibration using Validate calibration. To learn more see Validate Calibration.